Answer:
Part A with the effect of voltmeter = 16.15 V
Part B without the effect of voltmeter = 20.43 V
Part C Error = 20.94 %
Explanation:
Part A: What potential difference does the voltmeter measure across the 4.50kΩ ohm resistor?
Since voltmeter has its own internal resistance, when it measures voltage across an element then that voltage is not true voltage across the element since the effect of voltmeter is also included. The internal resistance of the voltmeter becomes parallel to the resistance of element
Req = Rvm*R/Rvm+R
Where Rvm is the resistance of voltmeter and R is the resistance of the element R = 4.50 kΩ
Req = 4.50*10/4.50+10
Req = 3.103 kΩ
Now we calculate the current flowing in the circuit since it is series circuit, the current is same through all elements
I = V/Rtotal
Where Rtotal = 6.50 + 3.103 = 9.603 kΩ
I = 50/9.603
I = 5.206 mA
V = I*Req
V = 5.206*3.103
V = 16.15 V
Part B: What is the true potential difference across the resistor when the meter is not present?
Now we don't want to include the effect of internal resistance of the voltmeter so
I = V/Rtotal
Now Rtotal = 6.50 + 4.50 = 11 kΩ
I = 50/11
I = 4.54 mA
V = I*R
V = 4.54*4.50
V = 20.43 V
Hence, as you can see the true voltage across the resistor is 20.43 V so internal resistance of the voltmeter shows reduced voltage across the resistor that was 16.15 V
Part C: By what percentage is the voltmeter reading in error from the true potential difference?
Percentage error = (20.43 - 16.15/16.15)*100 %
Percentage error = 20.94 %