Answer:
Explanation:
1) Resolution and uncertainty of both ADC ranges
a) Resolution (for +/- 5 mV range) = 5 mV / 2n where n = number of bits of ADC
Resolution = 5 mV / 28 = 5 mV / 256 = 0.0195 mV ..................1
uncertainty (for +/- 5 mV range) = +/- 0.5*resolution = +/- 0.5*0.0195 mV = 0.00975 mV ..........2
b) Resolution (for +/- 5 V range) = 5 V / 2n where n = number of bits of ADC
Resolution = 5 mV / 28 = 5 V / 256 = 0.0195 V = 19.5 mV ..................3
uncertainty (for +/- 5 mV range) = +/- 0.5*resolution = +/- 0.5*19.5 mV = 9.75 mV ..........4
2)Thermocouple sensitivity = ( Maximum output voltage - Minimum Output voltage) / (Maximum Temperature - Minimum Temperature)
Thermocouple sensitivity = (3.649mv - 0 ) / (70 - 0) = 0.0521 mV / Deg.C ............5
This is the required Thermocouple sensitivity
3) Water bath temperature is given as 57 deg.C
Hence voltage read by Thermocouple = Sensitivity*57 = 0.0521*57 mV = 2.9697 mV ........6
4)We need to use ADC with a range of +/- 5 mV range as ADC with +/- 5 V range can not do measurement as it's resolution is higher than output voltage.
ADC will measure voltage as 2.9695 mV ......................7
Oop
——————————————————————
Answer:A technology expert
Explanation:
Answer:
true
Explanation:
if it is not true it is false
Answer:
Please find the answer in the explanation
Explanation:
An employee of a communications company under the jurisdiction of the Public Utilities Commission when prepáring plans for communication lines can be exempted from the licensing provisions of the Professional Engineers Act.
Being an employee of a communications company under the jurisdiction of the Public Utilities Commission is not a criterion for getting a license to be a professional engineer nor a prerequisite for getting the license to be a professional engineer